analog scope problem
lightlinked
Tue Mar 23 2010, 03:42PM Print
lightlinked Registered Member #2087 Joined: Tue Apr 21 2009, 08:32AM
This issue has arisen with my analog scope. A lil love tap used to fix it but now it seems to persist. The worst part is that it affects the triggering. I dont even know how to explain it so here are some pictures. Nothing looks too amiss inside. Any ideas on what the problem is?
x/y mode no signal
1269358775 2087 FT0 Dsc 0731
no signal
1269358775 2087 FT0 Dsc 0735
with built in test signal
1269358775 2087 FT0 Dsc 0740
MinorityCarrier
Tue Mar 23 2010, 04:27PM
Since tapping it made it go away, that makes it a contact issue.
I like to gently tug on wire harnesses, gently rapping selector switches, and wiggling or re-seating edge connectors, to see if such problems can be made to disappear.
It depends on your comfort level on how invasive you want to be during trouble-shooting.
radiotech
Tue Mar 23 2010, 05:15PM
I looks like the beam switching logic is not blanking, or intermittantly blanking the beam as it shifts up and down.
You have 2 channels but only 1 pair of vertical plates and 1 vertical power amplifier, so a level swither is moving the beam between the places on the screen you set with the position pots.
The deflection amps in scopes like that one uses transistors and resistors that get hot and may crystalize the solder. Look for discolored resistors and cold solder joints.
In the 'no signal picture' the 'fade out' in the middle of the trace shows power filtering may be bad. With the sweep set low enought to produce that many square waves of the 1 kHz. probe calibation signal, a bad filter could get right down to crt blanking levels as well as cause much assorted mischief.
Whacking scopes like that isn't a good idea, it can dislodge structures inside the crt causing shorts, and a delicate reverse-whack
is needed to un-short them.
lightlinked
Wed Mar 24 2010, 01:45AM
the anomaly IS in sync with the mains... I didn't see any puffy caps or anything that looked burnt. Is it okay to scope the scope's own 12v and 5v power supply rails?
radiotech
Wed Mar 24 2010, 04:38AM
Check to see if heater of CRT is shorted to cathode.