Welcome
Username or Email:

Password:


Missing Code




[ ]
[ ]
Online
  • Guests: 37
  • Members: 0
  • Newest Member: omjtest
  • Most ever online: 396
    Guests: 396, Members: 0 on 12 Jan : 12:51
Members Birthdays:
All today's birthdays', congrats!
Download (31)
ScottH (37)


Next birthdays
11/03 Electroguy (94)
11/04 nitromarsjipan (2024)
11/04 mb (31)
Contact
If you need assistance, please send an email to forum at 4hv dot org. To ensure your email is not marked as spam, please include the phrase "4hv help" in the subject line. You can also find assistance via IRC, at irc.shadowworld.net, room #hvcomm.
Support 4hv.org!
Donate:
4hv.org is hosted on a dedicated server. Unfortunately, this server costs and we rely on the help of site members to keep 4hv.org running. Please consider donating. We will place your name on the thanks list and you'll be helping to keep 4hv.org alive and free for everyone. Members whose names appear in red bold have donated recently. Green bold denotes those who have recently donated to keep the server carbon neutral.


Special Thanks To:
  • Aaron Holmes
  • Aaron Wheeler
  • Adam Horden
  • Alan Scrimgeour
  • Andre
  • Andrew Haynes
  • Anonymous000
  • asabase
  • Austin Weil
  • barney
  • Barry
  • Bert Hickman
  • Bill Kukowski
  • Blitzorn
  • Brandon Paradelas
  • Bruce Bowling
  • BubeeMike
  • Byong Park
  • Cesiumsponge
  • Chris F.
  • Chris Hooper
  • Corey Worthington
  • Derek Woodroffe
  • Dalus
  • Dan Strother
  • Daniel Davis
  • Daniel Uhrenholt
  • datasheetarchive
  • Dave Billington
  • Dave Marshall
  • David F.
  • Dennis Rogers
  • drelectrix
  • Dr. John Gudenas
  • Dr. Spark
  • E.TexasTesla
  • eastvoltresearch
  • Eirik Taylor
  • Erik Dyakov
  • Erlend^SE
  • Finn Hammer
  • Firebug24k
  • GalliumMan
  • Gary Peterson
  • George Slade
  • GhostNull
  • Gordon Mcknight
  • Graham Armitage
  • Grant
  • GreySoul
  • Henry H
  • IamSmooth
  • In memory of Leo Powning
  • Jacob Cash
  • James Howells
  • James Pawson
  • Jeff Greenfield
  • Jeff Thomas
  • Jesse Frost
  • Jim Mitchell
  • jlr134
  • Joe Mastroianni
  • John Forcina
  • John Oberg
  • John Willcutt
  • Jon Newcomb
  • klugesmith
  • Leslie Wright
  • Lutz Hoffman
  • Mads Barnkob
  • Martin King
  • Mats Karlsson
  • Matt Gibson
  • Matthew Guidry
  • mbd
  • Michael D'Angelo
  • Mikkel
  • mileswaldron
  • mister_rf
  • Neil Foster
  • Nick de Smith
  • Nick Soroka
  • nicklenorp
  • Nik
  • Norman Stanley
  • Patrick Coleman
  • Paul Brodie
  • Paul Jordan
  • Paul Montgomery
  • Ped
  • Peter Krogen
  • Peter Terren
  • PhilGood
  • Richard Feldman
  • Robert Bush
  • Royce Bailey
  • Scott Fusare
  • Scott Newman
  • smiffy
  • Stella
  • Steven Busic
  • Steve Conner
  • Steve Jones
  • Steve Ward
  • Sulaiman
  • Thomas Coyle
  • Thomas A. Wallace
  • Thomas W
  • Timo
  • Torch
  • Ulf Jonsson
  • vasil
  • Vaxian
  • vladi mazzilli
  • wastehl
  • Weston
  • William Kim
  • William N.
  • William Stehl
  • Wesley Venis
The aforementioned have contributed financially to the continuing triumph of 4hv.org. They are deserving of my most heartfelt thanks.
Forums
4hv.org :: Forums :: General Science and Electronics
« Previous topic | Next topic »   

Checking the accuracy of domestic electricy meter...

1 2 
Move Thread LAN_403
Nicko
Tue Jan 12 2010, 05:19PM Print
Nicko Registered Member #1334 Joined: Tue Feb 19 2008, 04:37PM
Location: Nr. London, UK
Posts: 615
I'm in the UK. I suspect that our electricity utility provider's main meter is over-reading - its digital, 4 years old, 3-phase, 100A and has two time settings - an over-night rate that it switches to (called Economy 7 here) which is cheaper...

Now E.on (the provider) want to charge me GBP 75 to install a "check meter" in series - the problem is that I don't really want to spend the money as UK regulations say a meter is valid if it over-reads by 2.5% OR UNDER-READS by 3.5%. i.e. I could end up even more out of pocket.

How can I check this meter myself on the QT? Accurately. e.g. to 1 or 2%

PS. Individually checking each phase as well as the overall totals would be great.

Thanks
Back to top
radiotech
Tue Jan 12 2010, 08:43PM
radiotech Registered Member #2463 Joined: Wed Nov 11 2009, 03:49AM
Location:
Posts: 1546
First, check the creep. Does meter register anything with everything
loadside switched off. To check the meter you need a known load connected at the metersocket. Remember you pay for heating the copper from the meter to the loads you use. With known load and timer, you can check agreement. Why wont this be easy? (1) the kWhr meter may not have the same accuracy at all loads.(2) loads with low power factor, what ever it is, cost you the heating of the copper on your side of the meter to the load. The last suggestion is to try and get hold of the standard test method used to certify that meter for commerce.

Quite a while ago, when updating wiring, I found that a meter had been wired incorrectly in a gutterbox. This meant that one business tennant had been paying for 1/2 of another tennants power, although that tennant had paid full measure. This was subsequently repaired and the electric company refunded a lot of money. The records showed, for years the previous tennant had complained and the file for that address had letters unresolved due to language difficulties some 20 years previous.
Back to top
Nicko
Tue Jan 12 2010, 08:53PM
Nicko Registered Member #1334 Joined: Tue Feb 19 2008, 04:37PM
Location: Nr. London, UK
Posts: 615
We had the house completely rebuilt in 2005 - all the white goods are new with good PF - main heating is a heat pump with a commercial PF-adjusted compressor. There are some low energy bulbs, but not many - most of the lighting is halogen.

Disconnecting the house is probably futile - its all new wiring and shutting everything off would be a pain (there is a LOT of tech in here). What I want is to, over a period of some days, check that the meter is really registering the right number of KWh at the right times.

One bizarre feature is that we seem to be using 75 - 80% of our KWh at night - nothing happens at night! The fridge/freezer is still there. The washing machine & dishwasher might run, but the heat pump just ticks along 24/7 and doesn't care. No lighting is on except blue leds throughout the skirting boards in the common areas for floor lighting (about 30W in total).

I'm puzzled about what is going on...

Back to top
Electra
Wed Jan 13 2010, 06:26AM
Electra Registered Member #816 Joined: Sun Jun 03 2007, 07:29PM
Location:
Posts: 156
Wouldn’t the most straightforward way be to obtain a 3phase meter and wire it as a secondary meter between your company’s meter and your fuse board?
The total units for all time zones together should add up to roughly the same, to within 2x the meter tolerance’s of course, but at least this will eliminate any major fault with your meter.

Not to state the obvious but I thought economy 7 was only of benefit for large loads at night like water and storage heating. Are you sure you’ve not got a heatstore or something topping up your heating in the cold weather without you knowing.

Finally a bit of a wild one, and someone here will prove me wrong I guess, if you have an unbalenced reactive load connected between 2 differnt phases. Won’t this cause a positive apparent power in one phase and a negative apparent power in another phase, With respect to neutral ? In a mechanical meter these cancel each other out as they are on the same shaft so the net reading is zero, could an electronic meter have some calculating errors that do not subtract without some small error.
Surely they must test these new meters extensively though.
Back to top
radiotech
Wed Jan 13 2010, 07:18PM
radiotech Registered Member #2463 Joined: Wed Nov 11 2009, 03:49AM
Location:
Posts: 1546
Wattmeters measure E*I*Cos (phaseangle). Pure reactive loads will not register any wattage(in theory) but will have current flowing. The actual reading will be the I^R loss in the wires connecting the load. A wattmater can be made into a varmeter (volt amps reactive) by adding a phase shifting transformer. Here is an interesting paper. Link2

Be wary of measuring reactive loads! It is easy to burn out a wattmeter in the shop, mentioned from bitter experience.

Also, with anything three-phase, take nothing for granted. Elevators always had a phase sequence relay because if somthing happened in the way the feeders were connected later (repairs to transformers or overhead lines) the car would move in the opposite direction, rendering the saftey switches in the shaftway inoperable.
Back to top
Dinges
Wed Jan 13 2010, 09:31PM
Dinges Registered Member #2511 Joined: Mon Dec 07 2009, 02:46AM
Location:
Posts: 36
As others have said - disconnect everything in the house (as simple as switching all breakers off) -> electricity meter should now show 0W power use and kWh indication shouldn't increase.

Now plug in a load with known power consumption - I like a 2kW infrared heating lamp. Plug it in with a small kilowatthour meter (energy meter, cost about 25E; known also as 'killawat' in the USA; IMO, every household should have one to monitor the energy consumption of appliances - it will quickly repay for itself in saved energy costs). The value that the plug-in energy meter shows should agree with what the utility meter shows.

If there's a large discrepancy, you may want to repeat with different loads and graph the difference. As the plug-in energy meter doesn't have calibration marks, the readings it yields carry very limited legal weight. But it provides a strong indication on whether you should get the utility company to install a calibrated test meter.

Should your utility meter really be off, I expect you will get your 75 GBP refunded.

Over here, getting your meter verified is without extra charges.

If you can't be bothered to disconnect everything in the house because there's too much high-tech (I don't get that one, but so be it -consider it a man-made power outage) - pay the 75 GBP to have it metered for you. Though they may actually have to (briefly) disconnect your house from the grid entirely anyway to put the meter in series.

It really can't get much simpler than this. You can either do the measuring yourself, or have it done for you (and pay for it, of course)

Peter.
Back to top
Dave47
Thu Jan 14 2010, 01:26PM
Dave47 Registered Member #84 Joined: Thu Feb 09 2006, 01:06PM
Location: Dallas, TX
Posts: 47
These guys claim they will have a UK version available this quarter

Link2

I've been wanting one, but haven't pulled the trigger, yet.

David
Back to top
Nicko
Thu Jan 14 2010, 08:01PM
Nicko Registered Member #1334 Joined: Tue Feb 19 2008, 04:37PM
Location: Nr. London, UK
Posts: 615
The problem with all these units is that they are a compromise - as they are all driven by a clip-on CT for each phase rather than being in-line, you have to set the line voltage and they all assume a PF of 1 - our line voltage varies between 230VAC and 250VAC (or thereabouts). I have tried the Efergy meters, and they are OK, but unless you have true RMS & PF-aware meter its all rather pointless - we have quite a few low-energy lights, which have a dreadful PF...

Having said that, Google PowerMeter looks cool, and I will almost certainly get a GPM enabled device in due course...

I'm coming round to the idea that the only solution is to get the utility company to install a check meter...

rats frown
Back to top
Avalanche
Thu Jan 14 2010, 08:51PM
Avalanche Registered Member #103 Joined: Thu Feb 09 2006, 08:16PM
Location: Derby, UK
Posts: 845
at work we have a three-phase power meter with rs232 interface for logging. It's a bit like a current clamp meter, with voltage sense. calculates real and reactive power flow, quite a handy piece of kit.
Back to top
Dave47
Thu Jan 14 2010, 09:38PM
Dave47 Registered Member #84 Joined: Thu Feb 09 2006, 01:06PM
Location: Dallas, TX
Posts: 47
Nicko wrote ...

The problem with all these units is that they are a compromise - as they are all driven by a clip-on CT for each phase rather than being in-line, you have to set the line voltage and they all assume a PF of 1 - our line voltage varies between 230VAC and 250VAC (or thereabouts).

The TED unit looks like it also measures the voltage on both phases independently. That is, the box that is installed in the breaker panel has the two clip ons, but also has a voltage connection to each phase through a breaker.

Link2

Their website is a little light, so I didn't see where they specifically say it is correcting for PF, but it seems so. Anyway, they only have the US version at the moment, so it doesn't help you.

David
Back to top
1 2 

Moderator(s): Chris Russell, Noelle, Alex, Tesladownunder, Dave Marshall, Dave Billington, Bjørn, Steve Conner, Wolfram, Kizmo, Mads Barnkob

Go to:

Powered by e107 Forum System
 
Legal Information
This site is powered by e107, which is released under the GNU GPL License. All work on this site, except where otherwise noted, is licensed under a Creative Commons Attribution-ShareAlike 2.5 License. By submitting any information to this site, you agree that anything submitted will be so licensed. Please read our Disclaimer and Policies page for information on your rights and responsibilities regarding this site.