If you need assistance, please send an email to forum at 4hv dot org. To ensure your email is not marked as spam, please include the phrase "4hv help" in the subject line. You can also find assistance via IRC, at irc.shadowworld.net, room #hvcomm.
Support 4hv.org!
Donate:
4hv.org is hosted on a dedicated server. Unfortunately, this server costs and we rely on the help of site members to keep 4hv.org running. Please consider donating. We will place your name on the thanks list and you'll be helping to keep 4hv.org alive and free for everyone. Members whose names appear in red bold have donated recently. Green bold denotes those who have recently donated to keep the server carbon neutral.
Special Thanks To:
Aaron Holmes
Aaron Wheeler
Adam Horden
Alan Scrimgeour
Andre
Andrew Haynes
Anonymous000
asabase
Austin Weil
barney
Barry
Bert Hickman
Bill Kukowski
Blitzorn
Brandon Paradelas
Bruce Bowling
BubeeMike
Byong Park
Cesiumsponge
Chris F.
Chris Hooper
Corey Worthington
Derek Woodroffe
Dalus
Dan Strother
Daniel Davis
Daniel Uhrenholt
datasheetarchive
Dave Billington
Dave Marshall
David F.
Dennis Rogers
drelectrix
Dr. John Gudenas
Dr. Spark
E.TexasTesla
eastvoltresearch
Eirik Taylor
Erik Dyakov
Erlend^SE
Finn Hammer
Firebug24k
GalliumMan
Gary Peterson
George Slade
GhostNull
Gordon Mcknight
Graham Armitage
Grant
GreySoul
Henry H
IamSmooth
In memory of Leo Powning
Jacob Cash
James Howells
James Pawson
Jeff Greenfield
Jeff Thomas
Jesse Frost
Jim Mitchell
jlr134
Joe Mastroianni
John Forcina
John Oberg
John Willcutt
Jon Newcomb
klugesmith
Leslie Wright
Lutz Hoffman
Mads Barnkob
Martin King
Mats Karlsson
Matt Gibson
Matthew Guidry
mbd
Michael D'Angelo
Mikkel
mileswaldron
mister_rf
Neil Foster
Nick de Smith
Nick Soroka
nicklenorp
Nik
Norman Stanley
Patrick Coleman
Paul Brodie
Paul Jordan
Paul Montgomery
Ped
Peter Krogen
Peter Terren
PhilGood
Richard Feldman
Robert Bush
Royce Bailey
Scott Fusare
Scott Newman
smiffy
Stella
Steven Busic
Steve Conner
Steve Jones
Steve Ward
Sulaiman
Thomas Coyle
Thomas A. Wallace
Thomas W
Timo
Torch
Ulf Jonsson
vasil
Vaxian
vladi mazzilli
wastehl
Weston
William Kim
William N.
William Stehl
Wesley Venis
The aforementioned have contributed financially to the continuing triumph of 4hv.org. They are deserving of my most heartfelt thanks.
Registered Member #543
Joined: Tue Feb 20 2007, 04:26PM
Location: UK
Posts: 4992
James wrote ...
Get some lead sheet, 1mm thick should be plenty adequate for 50 kVp x-rays.
If we take operating conditions Va 50 kV, Ia 4 mA, filter Al 1mm, the dose rate at a distance of 50 cm will be 5 - 6 Gy/hr depending on beam geometry etc.
The addition of Pb shielding 1 mm will reduce the dose rate at 50 cm to roughly 50 mGy/hr - an unaccceptably high figure.
Registered Member #33
Joined: Sat Feb 04 2006, 01:31PM
Location: Norway
Posts: 971
I don't think the MacGyver method will work very well. X-ray developer and fixer should work just fine for developing photographic paper, as both x-ray film and photographic paper are based on the same chemistry.
Edit: I did some googling, and it turns out developing with orange juice is actually possible, and coffe supposedly works even better. It's doesn't work nearly as well as proper developer, of course, but it's actually possible.
Photographic paper is going to have a very low ISO, so it's not practical for direct exposures with a dental x-ray machine (or with a 2X2), you need an intensifying screen unless you want an exposure time of many hours. Ideally you would want blue-emitting x-ray screens, as I don't think photographic paper is sensitized to green light.
Proud Mary wrote ...
James wrote ...
Get some lead sheet, 1mm thick should be plenty adequate for 50 kVp x-rays.
If we take operating conditions Va 50 kV, Ia 4 mA, filter Al 1mm, the dose rate at a distance of 50 cm will be 5 - 6 Gy/hr depending on beam geometry etc.
The addition of Pb shielding 1 mm will reduce the dose rate at 50 cm to roughly 50 mGy/hr - an unaccceptably high figure.
Where are these figures from? They sound unrealistically high.
Registered Member #543
Joined: Tue Feb 20 2007, 04:26PM
Location: UK
Posts: 4992
Anders M. wrote ...
I don't think the MacGyver method will work very well. X-ray developer and fixer should work just fine for developing photographic paper, as both x-ray film and photographic paper are based on the same chemistry.
Photographic paper is going to have a very low ISO, so it's not practical for direct exposures with a dental x-ray machine (or with a 2X2), you need an intensifying screen unless you want an exposure time of many hours. Ideally you would want blue-emitting x-ray screens, as I don't think photographic paper is sensitized to green light.
Proud Mary wrote ...
James wrote ...
Get some lead sheet, 1mm thick should be plenty adequate for 50 kVp x-rays.
If we take operating conditions Va 50 kV, Ia 4 mA, filter Al 1mm, the dose rate at a distance of 50 cm will be 5 - 6 Gy/hr depending on beam geometry etc.
The addition of Pb shielding 1 mm will reduce the dose rate at 50 cm to roughly 50 mGy/hr - an unaccceptably high figure.
Where are these figures from? They sound unrealistically high.
I used RadPro, Anders and then verified its estimate against other sources. As you know, estimates of this kind are never going to be especially accurate, and are designed to err on the side of caution, but they serve very well to give an idea of the order of hazards involved.
Registered Member #33
Joined: Sat Feb 04 2006, 01:31PM
Location: Norway
Posts: 971
Proud Mary wrote ...
Anders M. wrote ...
I don't think the MacGyver method will work very well. X-ray developer and fixer should work just fine for developing photographic paper, as both x-ray film and photographic paper are based on the same chemistry.
Photographic paper is going to have a very low ISO, so it's not practical for direct exposures with a dental x-ray machine (or with a 2X2), you need an intensifying screen unless you want an exposure time of many hours. Ideally you would want blue-emitting x-ray screens, as I don't think photographic paper is sensitized to green light.
Proud Mary wrote ...
James wrote ...
Get some lead sheet, 1mm thick should be plenty adequate for 50 kVp x-rays.
If we take operating conditions Va 50 kV, Ia 4 mA, filter Al 1mm, the dose rate at a distance of 50 cm will be 5 - 6 Gy/hr depending on beam geometry etc.
The addition of Pb shielding 1 mm will reduce the dose rate at 50 cm to roughly 50 mGy/hr - an unaccceptably high figure.
Where are these figures from? They sound unrealistically high.
I used RadPro, Anders and then verified its estimate against other sources. As you know, estimates of this kind are never going to be especially accurate, and are designed to err on the side of caution, but they serve very well to give an idea of the order of hazards involved.
Your numbers suggest that the radiation from an x-ray tube operating at 50kV is attenuated two orders of magnitude by 1mm of lead, but this graph that you posted earlier states that even half a millimeter of lead will attenuate the x-rays by five orders of magnitude.
Registered Member #543
Joined: Tue Feb 20 2007, 04:26PM
Location: UK
Posts: 4992
Anders M. wrote ...
Your numbers suggest that the radiation from an x-ray tube operating at 50kV is attenuated two orders of magnitude by 1mm of lead, but this graph that you posted earlier states that even half a millimeter of lead will attenuate the x-rays by five orders of magnitude.
I don't see any contradiction here. The figures in the graph are derived from probablistic energy-dependent half-values without reference to fluence, whilst the figures I have proposed are for dose rates in Gy/hr 50 cm from the anode at beam cone centre.
Registered Member #33
Joined: Sat Feb 04 2006, 01:31PM
Location: Norway
Posts: 971
What is the practical difference between those two figures, why is there such a large difference between them, and why are the numbers from RadPro more applicable? I'm just trying to learn more about this, and this seems a bit counterintuitive to me.
Registered Member #1412
Joined: Thu Mar 27 2008, 04:07PM
Location: Taipei Taiwan
Posts: 278
I test MacGyver's method.But I use some overdue vitamin pills and ammonia.It works but ofcourse very low quality.(not with X-ray,With ordinary negative film)
This is the very first time I develop photo.
But my dad was VERY VERY angry about the smell of ammonia.
Registered Member #543
Joined: Tue Feb 20 2007, 04:26PM
Location: UK
Posts: 4992
Anders M. wrote ...
What is the practical difference between those two figures, why is there such a large difference between them, and why are the numbers from RadPro more applicable? I'm just trying to learn more about this, and this seems a bit counterintuitive to me.
Here is a paper by Ray McGinnis, designer of RadPro, on some of the assumptions he used in the calculator design:
Registered Member #33
Joined: Sat Feb 04 2006, 01:31PM
Location: Norway
Posts: 971
Yes, I've seen reports of people using NaCO3 together with orange juice, so this should work fine. Here is some more info about developing film with coffee.
Photographic paper is sensitive to x-rays, but it's less sensitive to x-rays (and light) than film, because paper usually has a much lower ISO rating (5-25 according to some forum posts I found) than film (100-3000 usually).
Proud Mary wrote ...
Anders M. wrote ...
What is the practical difference between those two figures, why is there such a large difference between them, and why are the numbers from RadPro more applicable? I'm just trying to learn more about this, and this seems a bit counterintuitive to me.
Here is a paper by Ray McGinnis, designer of RadPro, on some of the assumptions he used in the calculator design:
Shielding Equations and Buildup Factors Explained
I would think that the numbers from the NIST graph would me more applicable in this case, as these are the guidlines used for x-ray shielding in practice. The numbers from RadPro are based on theoretical calculations with a lot of unrealistic assumptions. To quote from the RadPro website: "It should always overestimate the amount of shielding required because e-max for the tube head is used when in reality the device produces an entire energy spectrum from lower energies up to e-max.". I think overestimate is a mild word in this case, as assuming all of the energy is at e-max is quite a bold assumption.
This site is powered by e107, which is released under the GNU GPL License. All work on this site, except where otherwise noted, is licensed under a Creative Commons Attribution-ShareAlike 2.5 License. By submitting any information to this site, you agree that anything submitted will be so licensed. Please read our Disclaimer and Policies page for information on your rights and responsibilities regarding this site.