Help with a HV source for X-Rays

Neukyhm, Thu May 11 2017, 04:59PM

Hi again, 4hv users. I need help with a 70KV power supply I will build soon. The problem is that the cathode goes to the heated filament of the tube so x-rays are made. If I do this, I will have to build a small isolation transformer for the filament that runs with 2.5-3V. I will also have to isolate the Cockroft Walton multiplier and tube anode with epoxy, see diagram:

(Note: the multiplier cathode goes from the beginning of the multiplier of course, I have drawn it wrong, corrected in next diagram)

1494521278 57520 FT0 Fullsizerender


(Note: that resistor between filament and grid is required so emision won't be very high)

But I asked for advice to the tube manufacturer and they said that if I ground the cathode, I will only have to isolate the multiplier and anode. This would be great because it means that I would be able to run the filament with any DC source, isolated or not, because the filament is connected to the cathode and this would be at 0V. See picture where I have added grounds.

1494521472 57520 FT0 Fullsizerender 1


I have experience in HV and I have already built a X-Ray machine, but this "grounding method" is new to me, I don't even know if those grounds are well placed. So 4HV users that are more experienced, please give me any advice you have.
Re: Help with a HV source for X-Rays
Sulaiman, Thu May 11 2017, 05:51PM

the thing to consider is anode cooling if at eht
Re: Help with a HV source for X-Rays
Neukyhm, Thu May 11 2017, 06:44PM

Sulaiman wrote ...

the thing to consider is anode cooling if at eht
No idea what you mean with "eht", english is not my first language.
Re: Help with a HV source for X-Rays
johnf, Thu May 11 2017, 07:00PM

EHT = extra high tension = extra high voltage
Re: Help with a HV source for X-Rays
teravolt, Thu May 11 2017, 08:07PM

have you thought about using a small marx to give pulsed ex-rays
Re: Help with a HV source for X-Rays
Neukyhm, Thu May 11 2017, 08:44PM

teravolt wrote ...

have you thought about using a small marx to give pulsed ex-rays
Yes but I'm not looking for pulsed x-rays. Btw the tube is a CEI OX/70-G4 and I will use it at about 120W, I don't expect the cooling to be a problem here.

I want to build a C-W multiplier like CEI has: Link2
but I'm not sure how I have to connect grounds, I mean, that square wave is coming from arduino, is it safe to connect arduino, primary coil and C-W multiplier grounds all together? This is the way that CEI uses, and it is the best since you don't need insulation on cathode's side because it's grounded and at 0 volts.
Re: Help with a HV source for X-Rays
ScotchTapeLord, Mon May 15 2017, 11:38AM

What Sulaiman was talking about was having to float your anode cooling system at high voltage. This can be more complicated than floating a filament. How do you plan to cool your anode?
Re: Help with a HV source for X-Rays
Proud Mary, Mon May 15 2017, 07:12PM

Rechargeable batteries are a very simple, practical, and reliable, way of implementing a floating heater or filament supply.

You would in any case want to start the heaters in a standby mode before applying EHT to avoid damaging the tube, so you can simply switch your batteries into the heater with a little local switch - with battery and switch close to the tube - before you start the High Voltage supply.

If you want to control the heater current as a means of controlling the tube current, you can just use a high wattage variable resistor, one of the big ceramic ones, or an old fashioned rheostat.

Since most tubes need filament supplies between 1V - 4V you can also make this up out of combinations of rechargeable cells, for example:

1.2V - NiCd
2V - Pb-acid
2.4V - 2*NiCd
3.2V NiCd + Pb-acid
3.6V 3*NiCd or Li
4V - 2*Pb-acid

Take care you choose cells with sufficient Ah capacity, as your tube current will start to fall with the filament voltage, so the anode voltage will rise (unless regulated) with a risk of flash over - also if the filament voltage falls, causing the anode voltage to rise, the tube will produce fewer but harder rays. If the filament voltage fails completely, the tube may go into field emission mode, losing focus, and emitting x-rays in unpredictable directions.