Internals of a Tek P6013A HV O-scope probe.

Patrick, Wed Feb 02 2011, 05:29AM

ok i start with pics.
1296624584 2431 FT0 Sam 0302resized
The total device (As bougt from Ebay).

1296624584 2431 FT0 Sam 0301uhfcropped
Radio Shack UHF to BNC changer.

1296624584 2431 FT0 Sam 0304cropped

1296624584 2431 FT0 Sam 0305resiszed

1296624584 2431 FT0 Sam 0307cropresize

1296624584 2431 FT0 Sam 0309resized

1296624584 2431 FT0 Sam 0311resize

1296624584 2431 FT0 Sam 0321leafr

1296624584 2431 FT0 Sam 0323text

1296624584 2431 FT0 Sam 0324text

1296624584 2431 FT0 Sam 0329text

1296624584 2431 FT0 Sam 0333solder

1296624584 2431 FT0 Sam 0336textre
Re: Internals of a Tek P6013A HV O-scope probe.
Steve Maurer, Fri Feb 04 2011, 04:25PM

Thank you for sharing these pictures with us.

I find the leaf capacitor arrangement interesting.

- Steve
Re: Internals of a Tek P6013A HV O-scope probe.
Pinky's Brain, Fri Feb 04 2011, 04:45PM

I assume it's there to couple the high frequency components to the output? Because without it the grounded shield pulls the output signal to ground?
Re: Internals of a Tek P6013A HV O-scope probe.
Patrick, Fri Feb 04 2011, 06:08PM

Pinky's Brain wrote ...

I assume it's there to couple the high frequency components to the output? Because without it the grounded shield pulls the output signal to ground?
This statement eludes me, please clarify.
Re: Internals of a Tek P6013A HV O-scope probe.
Pinky's Brain, Sat Feb 05 2011, 12:58AM

Well the grounded shielding has a large parasitic capacitance to the resistor, if that was all there was to the probe then for higher frequencies most of the signal gets coupled to the shielding long before it gets to the low voltage arm of the divider (the electric field along the resistor is no longer symmetrical like it is for instance with your resistor from the other threads with the hoops). I assume the leaf capacitor is not grounded, but connects to the low voltage arm ... so I assume it's there to couple the high frequency components to the low voltage arm to counteract the effect from the shielding ...
Re: Internals of a Tek P6013A HV O-scope probe.
Patrick, Sat Feb 05 2011, 04:44AM

Pinky's Brain wrote ...

Well the grounded shielding has a large parasitic capacitance to the resistor, if that was all there was to the probe then for higher frequencies most of the signal gets coupled to the shielding long before it gets to the low voltage arm of the divider (the electric field along the resistor is no longer symmetrical like it is for instance with your resistor from the other threads with the hoops). I assume the leaf capacitor is not grounded, but connects to the low voltage arm ... so I assume it's there to couple the high frequency components to the low voltage arm to counteract the effect from the shielding ...
Do you mean the 1k (R111) hold up resistor...at HF the 1k (Common Cartbon Comp, 1/2 W) can get much hotter then the HV resistor! charging 3pf and discharging at 75Mhz and 200v must make a lot of heat.



1296881064 2431 FT107663 P6013a 1
the coax cable, the length from the probe can end, torwards the comp-box spans 10Ft and the center strand has 225K resistrence. therefore 225K/10ft = 22.5kohms per foot.


1296881064 2431 FT107663 P6013a 2

1296881064 2431 FT107663 P6013a 3
Re: Internals of a Tek P6013A HV O-scope probe.
Pinky's Brain, Sat Feb 05 2011, 02:49PM

Patrick wrote ...
Do you mean the 1k (R111) hold up resistor
With low voltage arm? I mean the 1K resistor and everything to the right of it in the schematic.

To me it seems that a handheld probe is significantly harder to design than a standalone probe connected through a low induction copper strap. With a stand alone probe you can keep the electric field across the HV resistor uniform (like your capacitive hoops in the other thread) which simplifies modelling it's operation.

The moment you put a grounded metal shield around it which goes close to the tip the behaviour of the system becomes much harder to predict.
Re: Internals of a Tek P6013A HV O-scope probe.
Patrick, Sat Feb 05 2011, 07:42PM

Pinky's Brain wrote ...

The moment you put a grounded metal shield around it which goes close to the tip the behaviour of the system becomes much harder to predict.
Yes I worry about this. The NorthStar's dont even have sheilding running up torwards the tip at all , just that blue nylon.
Re: Internals of a Tek P6013A HV O-scope probe.
Patrick, Mon Feb 07 2011, 05:37AM

Pinky's Brain wrote ...

To me it seems that a handheld probe is significantly harder to design than a standalone probe connected through a low induction copper strap. With a stand alone probe you can keep the electric field across the HV resistor uniform (like your capacitive hoops in the other thread) which simplifies modelling it's operation.

The moment you put a grounded metal shield around it which goes close to the tip the behaviour of the system becomes much harder to predict.
The NorthStar's do have a large topand bottom ring, to equalize the field and as you said I guess thats where some capacity comes from too.

I believe the Tek P6013, P6014, P6015 all have that metal shield so they can treat the whole path as a lossy-transmission line?

What about this...

1297057067 2431 FT107663 Hfhv1
Yellow are copper "Park-like" Caps, Red are resistors. The bottom most section are where compensation and LV componets live.


1297057067 2431 FT107663 Hfhv2



1297057067 2431 FT107663 Hvhf3
Re: Internals of a Tek P6013A HV O-scope probe.
Patrick, Sat Oct 01 2011, 11:51PM

Ok ive got my silver epoxy so ill make the attempt to fix this probe tonight.

Silver epoxy
1317512484 2431 FT1630 Silver Epoxy
Re: Internals of a Tek P6013A HV O-scope probe.
Patrick, Mon Oct 03 2011, 10:29PM

Ive come to the conclusion i may not be able to fix this probe. But I have still learned alot so it was worth the 60 USD and teh time lost trying to fix it.

First, the capacitence (3pf) at 60 hz was enough to damp the 10kV sine signal to just 28 mV! through a 99.8 Mohm resistor. And though the compensation box is hard to fix, when my body's stray capacitence was attached to the LV node, the O-scope showed a perfect 17V sine wave. so capacittance does play an enormous role in how these devices work.

I continue the my building of my own probes.