I am building a quarter shrinker. I plan on using an analog voltmeter to measure the charge on the capacitor. The meter I plan on using is an old Weston 2500 volt meter. It has a resistance of 85 ohms and a full-scale current draw of 1 ma. I want to set it up to measure a maximum of 25 KV. To do this I am using a multiplier resistor of 25M ohms. The resistor is a Caddock HV resistor rated for 30 KV and 15 watts.

I see some applications that use a voltage divider using 3 resistors: the meter internal resistance, another resistor in parallel with the meter resistance, and the multiplier resistor. Is this really necessary? I plan on using a voltage divider using only 2 resistors: the meter resistance and the multiplier resistor.

I am also wondering if I should ground the meter terminal without the multiplier resistor to earth to avoid the possibility of the meter floating up to an excessively high voltage.

The entire quarter shrinker will be completely isolated from the mains except during charging. After the final charge voltage has been reached, the charging connections to the capacitor will be physically disconnected.
Re: Measuring HV using an analog voltmeter
Patrick, Wed May 01 2019, 06:50PM

MRMILSTAR wrote ...

I plan on using a voltage divider using only 2 resistors: the meter resistance and the multiplier resistor.

dont do it. ill explain when i get home from work.

follow up: are you using the meter movement or the whole meter ?


Re: Measuring HV using an analog voltmeter
teravolt, Wed May 01 2019, 07:09PM

since your mesureing the charge voltage you will need a lower divider resistor add a varistor or transorb and a small capacitance like .1uf in parallel with it. the lower resisitance will be your meter resistance in parallel with your chosen resistance. remember that your coin crusher is ferly dangerouse and your meter should not be near your person and just where you can see it. maby there should be a lexan shield between you and your device. please treat with respect and use a crowbar or proper shorting stick with your caps, it could be deadly
Re: Measuring HV using an analog voltmeter
MRMILSTAR, Wed May 01 2019, 10:06PM

Patrick wrote ...

MRMILSTAR wrote ...

I plan on using a voltage divider using only 2 resistors: the meter resistance and the multiplier resistor.

dont do it. ill explain when i get home from work.

follow up: are you using the meter movement or the whole meter ?

I measured the resistance of the meter as 85 ohms with my Fluke multimeter. I'm not sure what you mean my "the whole meter" or "just the meter movement". This is not a multimeter. It is a Weston stand-alone dedicated panel-mount voltmeter. It has a full-scale reading of 2500 volts with a factory-specified external multiplier resistor. I don't know what the factory-specified external multiplier resistor is. I assume it is 2.5M ohms. With the factory-specified multiplier resistor, the sensitivity is specified as 1000 ohms/volt on the face plate. I am not using the factory-specified multiplier resistor. I am using my own Caddock external multiplier resistor value of 25M ohms.
Re: Measuring HV using an analog voltmeter
Patrick, Wed May 01 2019, 11:54PM

MRMILSTAR wrote ...

This is not a multimeter. It is a Weston stand-alone dedicated panel-mount voltmeter.

Yes I was hoping you meant this.




Re: Measuring HV using an analog voltmeter
Patrick, Thu May 02 2019, 06:35AM

Link2

here is a schematic...
Re: Measuring HV using an analog voltmeter
Patrick, Thu May 02 2019, 06:45AM

Patrick wrote ...

Link2

here is a schematic...


Re: Measuring HV using an analog voltmeter
Patrick, Thu May 02 2019, 06:45AM

Patrick wrote ...



here is a schematic...


Re: Measuring HV using an analog voltmeter
Patrick, Thu May 02 2019, 06:46AM

Patrick wrote ...

Patrick wrote ...


here is a schematic...


Re: Measuring HV using an analog voltmeter
Patrick, Thu May 02 2019, 06:46AM

Patrick wrote ...

Patrick wrote ...

Patrick wrote ...

Link2
here is a schematic...


Re: Measuring HV using an analog voltmeter
Sulaiman, Thu May 02 2019, 01:39PM

The meter is 1 mA full scale deflection.

If you use a 25 MOhm resistor in series then full scale deflection is equivalent to 25 kV
(25 kV x 1mA = 25W which is much more than the 15W rating that you mentioned)
Asuming that the resistor is exactly 25 MOhm
and that full scale deflection is exactly 1.000 mA

The voltage across the meter movement will be 85 mV at fsd.

For current meters an additional resistor is often added in series to give a convenient voltage,
e.g. 15 Ohms would give 100 mV total at 1 mA,
often this resistor will compensate for coil resistance changes with tenperature.

For a voltmeter application it was very difficult to make a meter exactly 1.00 mA fsd,
and the AlNiCo magnets would weaken over time so sensitivity reduced,
so the actual movement would be more sensitive than 1 mA fsd,
and a resistor was added across the movement to reduce the sensitivity / calibrate the meter.

... from unreliable memory



So, you may want to add a trimmable resistance in parallel with the movement,
or a trimmable resistance in series with the meter,
or be satisfied with the tolerances of the movement and resistor

If you can measure the fsd current sensitivity of the movement then you can add precisely the correct series resistance for accurate readings,
but plus or minus a few percent is usually good enough for most practical purposes.
Re: Measuring HV using an analog voltmeter
klugesmith, Thu May 02 2019, 03:41PM

Along the lines of what Sulaiman said, it's good to design in some adjustability if you want an instrument accurate to 5% or better. Of course that depends on having access to a high voltage reference for calibration, or accurately measuring the HV resistor and knowing about its voltage coefficient of resistance. Same considerations apply with digital panel meters.

Dialing down the sensitivity with a parallel resistor is easy. It adds safety, because there's a place for HV current to go if the meter broke & became open circuit.

Dialing down sensitivity with a series resistor needs to reduce the HV current, so the adjustment resistor must handle a lot of voltage and power.

Your 1 mA HV meter is going to slow down the capacitor charging, and start bleeding the charge down as soon as you disconnect the charger.
Why don't you find a nice pretty panel meter with a 100 microamp movement, or less, such as in analog HV measuring probes for CRT service?
Then your HV resistor will draw a lot less power, and be physically easier to manage.

The 4.5 x 4 inch analog panel meter in my hand has a scale card saying D.C. KILOVOLTS. Arc is labeled 0 to 25 on one side and 0 to 10 on the other. Fine print says FS = 50 uA.

If pictures were working, I would show my cap-charging control box that includes a Nixie-tube panel DVM and a HV resistor of hundreds of megohms. Also a high voltage switch. Switch positions are 1) connect charger; 2) disconnect charger, so only the meter is connected; 3) ground the HV wire, to bleed capacitor down. High power HV charging resistor is in series at the capacitor end of wire, so no failure of charger or control box can cause an explosion.

Capacitor voltage is accurately indicated whenever switch is in middle position (ready to fire). It can be bumped upward or downward by briefly turning the switch one way or the other. With your "floating" capacitor, you could get the same functionality and safety with charging resistors on both sides of the capacitor, two HV wires, and a double pole switch in meter/control box.
Re: Measuring HV using an analog voltmeter
MRMILSTAR, Thu May 02 2019, 03:48PM

Patrick wrote ...

Patrick wrote ...

Link2

here is a schematic...

Thanks. I was contemplating adding that earth ground to prevent possible damage to the meter. I will definitely add that earth ground.


Re: Measuring HV using an analog voltmeter
MRMILSTAR, Thu May 02 2019, 04:54PM

klugesmith wrote ...

Along the lines of what Sulaiman said, it's good to design in some adjustability if you want an instrument accurate to 5% or better. Of course that depends on having access to a high voltage reference for calibration, or accurately measuring the HV resistor and knowing about its voltage coefficient of resistance. Same considerations apply with digital panel meters.

Dialing down the sensitivity with a parallel resistor is easy. It adds safety, because there's a place for HV current to go if the meter broke & became open circuit.

Dialing down sensitivity with a series resistor needs to reduce the HV current, so the adjustment resistor must handle a lot of voltage and power.

Your 1 mA HV meter is going to slow down the capacitor charging, and start bleeding the charge down as soon as you disconnect the charger.
Why don't you find a nice pretty panel meter with a 100 microamp movement, or less, such as in analog HV measuring probes for CRT service?
Then your HV resistor will draw a lot less power, and be physically easier to manage.

The 4.5 x 4 inch analog panel meter in my hand has a scale card saying D.C. KILOVOLTS. Arc is labeled 0 to 25 on one side and 0 to 10 on the other. Fine print says FS = 50 uA.

If pictures were working, I would show my cap-charging control box that includes a Nixie-tube panel DVM and a HV resistor of hundreds of megohms. Also a high voltage switch. Switch positions are 1) connect charger; 2) disconnect charger, so only the meter is connected; 3) ground the HV wire, to bleed capacitor down. High power HV charging resistor is in series at the capacitor end of wire, so no failure of charger or control box can cause an explosion.

Capacitor voltage is accurately indicated whenever switch is in middle position (ready to fire). It can be bumped upward or downward by briefly turning the switch one way or the other. With your "floating" capacitor, you could get the same functionality and safety with charging resistors on both sides of the capacitor, two HV wires, and a double pole switch in meter/control box.

Here is a little more explanation of my design. There will be 2 volt meters. One will be for charging and the other will be for discharging into a bank of power resistors. The charging circuit and meter will be physically disconnected by a double-pole switch which will actually consist of pulling 2 charging clamps free via lanyards from a distance. The discharge meter will then be connected via a single-pole switch which will look like a mousetrap. The other side of the discharge voltmeter will be grounded. Everything will be done remotely via non-conductive lanyards. The charging circuit will be physically disconnected before firing.

Efficiency is not really a problem. I am using a 1 KVA NST (15 KV AC at 60 ma) to charge the capacitor. The charging resistor will consume less than 2% of the NST power. The maximum voltage that I will probably ever charge the capacitor with will be around 15 KV DC which corresponds to about 11,000 joules and 9 watts of power dissipation in the charging multiplier resistor.

High accuracy is not really a requirement. Within 1 KV is probably good enough for this application. If for no other reason is that I do not have access to a calibrated HV supply to set the meters more accurately with.

I was looking for 2 identical volt meters with 25 times some factor of 10 at full scale (25, 250, 2500, 25000) to make true voltage easy to read. It also makes the task of finding suitable HV power resistors simpler, at least for a reasonable price on E-Bay. 25M ohm, 15 watt , 30 KV Caddocks were readily available. Having a lower full scale current draw would have been nice but it was hard enough to find 2 identical volt meters without posing an additional requirement of low current at full scale. So I have to accommodate whatever full scale current draw my meters need which in this case is 1 ma.

I do have a question about voltmeter topology. Here are the 2 options:

1. Use a single 25M ohm multiplier resistor on each voltmeter with the other leg grounded.
2. Use 2 multiplier resistors per voltmeter, one per leg with the total adding up to 25M ohms, and no ground?

Re: Measuring HV using an analog voltmeter
Patrick, Thu May 02 2019, 05:07PM

MRMILSTAR wrote ...

Patrick wrote ...

Patrick wrote ...

Link2

here is a schematic...

Thanks. I was contemplating adding that earth ground to prevent possible damage to the meter. I will definitely add that earth ground.
yes you can peruse my imgur page for anything not posted here on our forum.

ill be using the 5600 volt Micsig oscilloscope probes here soon too.

Re: Measuring HV using an analog voltmeter
klugesmith, Thu May 02 2019, 06:22PM

MRMILSTAR wrote ...
... There will be 2 volt meters. One will be for charging and the other will be for discharging into a bank of power resistors. The charging circuit and meter will be physically disconnected by a double-pole switch which will actually consist of pulling 2 charging clamps free via lanyards from a distance. The discharge meter will then be connected via a single-pole switch which will look like a mousetrap. The other side of the discharge voltmeter will be grounded. Everything will be done remotely via non-conductive lanyards. The charging circuit will be physically disconnected before firing.

OK. So after a shot, you won't know how much charge is left in the capacitor, until you close the discharge switch and the voltage immediately begins to drop. If the quarter-shrinking circuit is underdamped, and doesn't go open circuit until the second or third quarter-cycle, the residual voltage on capacitor could even be negative. A popular time for arcs to go out is zero-current points, for example at the end of second quarter-cycle, with maximum voltage reversal on the capacitor. What do your quarter-shrinking references say about residual voltage?

Between disconnecting charger and firing a shot, you won't know how much charge has been lost through leakage or corona. If charging resistor is on capacitor side of the meter, meter will read high while charging current flows. (not a problem if you terminate charging by backing off on a variac that feeds the NST). If charging resistor is on charger side of the meter, there's risk of an explosive discharge from any sparking on capacitor side of the resistor.

Here is a quarter shrinker accident report I had no trouble re-finding today:
Link2 Science/Quarter Shrinker/Accident/accident.htm

Here's one more idea, while I respect that it's _your_ project. Suppose you have just one meter, so no need to scrounge a matched pair. Set it up next to the capacitor, always connected. If it's too hard to read from a safe distance, use a tripod-mounted spotting scope or binoculars. Make it 0.10 mA or 0.05 mA, like most kV meters, to reduce the bleed rate after you disconnect the charger. Charge a bit higher than you want, disconnect charger, then wait for bleed down to desired target voltage.

Back to your question about one vs two HV metering resistors. If the meter(s) is at operator's end of HV cables, at what voltage will its case and terminals be with respect to earth ground? I had one experiment with a meter floating at some 5-figure voltage, where I think the misbehavior was related to electrostatic forces upon delicate moving parts.
Re: Measuring HV using an analog voltmeter
MRMILSTAR, Fri May 03 2019, 04:12AM

klugesmith -

Thanks for the replies. I was thinking that I shouldn't leave the meter connected during a discharge because I was afraid that the discharge would somehow damage the meter. If the meter will not be damaged during the discharge then perhaps I can use just a single meter permanently connected. I can find a lower current meter if I don't need to find two identical ones. Do you use a single meter permanently connected to the capacitor? Do you ground one side of the meter? Grounding seems like a good idea to protect the meter.

Just to clear up a misunderstanding, the intention was never to have a meter near the operator. It would be co-located with the rest of the apparatus on a cart. Binoculars will be used if needed to read the meter. The only thing near the operator will be the variac.

I was planning on using a remote variac to charge the capacitor. When the charge level has been reached, I was just going to quickly disconnect the charging cables at full voltage via lanyards without turning down the variac. No switch will be involved, just 2 jumper cables that can quickly be removed by lanyards. My reasoning is that if I turn down the variac to zero after charging, the only thing holding back the capacitor voltage from feeding into the NST output is the full wave bridge rectifier. If I have a diode failure at that time, the full massive current from the capacitor could potentially go into the NST output terminals and then, via an internal short in the NST, possibly into the power cord leading back to the variac where I will be located. Not a good scenario.

What is your procedure during charging on your quarter shrinker?
Re: Measuring HV using an analog voltmeter
klugesmith, Fri May 03 2019, 06:31AM

Isn't Signification a regular on this forum, who did some quarter shrinking?
My own electroforming work never got past can crushing and can cutting-in-half, in about 2007, using up to 1 kJ in a capacitor whose ratings support 10+ kJ.

Your question inspired me to draw a schematic & try to post it here. Guess it needs clicking.
Link2

The HV switch poles are far enough apart, and the moving contact arm is long enough, that arcs don't persist when you open the charging or discharging path. The only way you could ever have more an ampere of current in the wires between switch and capacitor, is if both charging resistors shorted. They're ceramic tubes about 7 inches long and 1 inch in diameter.
Re: Measuring HV using an analog voltmeter
klugesmith, Tue May 07 2019, 04:05AM

Oops, that link I posted to a quarter shrinker accident report & failure analysis was broken.
Let's try again. I think the root of all evil is URL's with white space in the middle, as if there were any good reason to allow it.

Again it didn't work. Before I edited this, I think the automatic link was OK in preview pane, and OK after hitting Post Reply button, and then went bad when I used "edit" for some unrelated wording change.

Cutting and pasting the follwing URL ought to work.
"http://www.brianb.org/images/Backyard Science/Quarter Shrinker/Accident/accident.htm"


Re: Measuring HV using an analog voltmeter
klugesmith, Tue May 07 2019, 04:10AM

Didn't we used to have a way to delete our own posts, before there were any replies?
Re: Measuring HV using an analog voltmeter
radiotech, Wed May 08 2019, 10:01AM

In some cases damping of a meter movement is desired,
and connecting a resistor across the movement accomplishes
this. For a meter connected in series with a large series multiplier,
as far as the meter is concerned, it is being fed from a constant
current source.

If you tell me the Weston meter series number, I may have
the catalog data, if it is a very old one. Since you have a low
resistance movement , you might consider a MOV rated at
a hundred or so volts. The resistance will not affect the meter,
but if the meter movement opens, it will prevent arking that
might ignite something.

I have no idea how many joules your capacitor will store,
but if lethal, something more substantial than that meter
should be in place to warn it is charged.
Re: Measuring HV using an analog voltmeter
MRMILSTAR, Wed May 08 2019, 02:30PM

I read about Brian's accident and have studied his design carefully. Three things made me uncomfortable with his implementation.

1. Everything seems too crowded for the voltage and power levels.
2. He used a metal cart.
3. The circuitry just seems overly complicated and relies too much on electronics for safety. Protection is good but I think there are simpler and more reliable ways to accomplish this.

All of these things just seems to invite flash-overs.

Before I started my design I studied all of the quarter shrinkers for which I could find information. I believe that I am incorporating the best features of these and adding some of my own. For example, in my design, all high voltage equipment and wiring is widely spaced and I am using a solid plastic cart to avoid flash-overs. I am also relying strictly on mechanical means activated by long non-conductive lanyards to trigger events. There will be no electronics to invite failure.
Re: Measuring HV using an analog voltmeter
MRMILSTAR, Wed May 08 2019, 06:39PM

radiotech wrote ...

In some cases damping of a meter movement is desired,
and connecting a resistor across the movement accomplishes
this. For a meter connected in series with a large series multiplier,
as far as the meter is concerned, it is being fed from a constant
current source.

If you tell me the Weston meter series number, I may have
the catalog data, if it is a very old one. Since you have a low
resistance movement , you might consider a MOV rated at
a hundred or so volts. The resistance will not affect the meter,
but if the meter movement opens, it will prevent arking that
might ignite something.

I have no idea how many joules your capacitor will store,
but if lethal, something more substantial than that meter
should be in place to warn it is charged.

Because the Weston meter requires so much current (1 ma) and thus being quite inefficient, I have decided to use 2 Simpson microammeters instead that each require only 25 microamps full scale. A HV 1G ohm (two HV 500G ohm resistors in series) multiplier resistor will be used for each one. One meter will be used for charging. It reads 0-25 microamps, so I just multiply the scale by 1000 to get the charging voltage. It will be disconnected after charging and prior to firing. The other meter will be used for resistor bleed-off. Its scale is 25-0-25 microamps to account for a potential negative residual capacitor charge caused by voltage reversals during firing. It will be disconnected during charging and firing and connected during resistor bleed-off.

The ultimate final check for residual charge will be a chicken stick. I do not trust electronics for this.

If fully charged to its maximum rated voltage of 22 KV, the energy would be 24K joules. I never expect to charge it to that level because that is much more than is needed and would be quite stressful on the capacitor due to excessive voltage reversals. To maximize capacitor life, I will restrict the charge to about 15 KV which will result in 11,250 joules.

If you do have access to a spec sheet for the Weston voltmeter I would still like to have it for future reference. All that I know about it is what is printed on the face plate as follows:

* Made in 1958 (stamped on rear of case)
* Model 741
* 0 - 2500 volts full scale
* FS = 1 ma (presumably with external multiplier resistor #9492)
* 1000 ohms/volt (presumably with external multiplier resistor #9492)
* Use with external resistor #9492

The measured resistance with my Fluke DMM is 85 ohms. I don't know what value the external resistor is but I assume that it is 2.5M ohms based on the other information.

Re: Measuring HV using an analog voltmeter
radiotech, Thu May 09 2019, 09:39AM

Here is data on the Weston 741 series and multiplier resistors. It is dated 1955.

Re: Measuring HV using an analog voltmeter
MRMILSTAR, Thu May 09 2019, 03:19PM

The Weston info didn't show up. I sure wish we could get this image insertion problem with the web site fixed. The ability to attach pictures is one of the reasons that I really liked this forum.
Re: Measuring HV using an analog voltmeter
radiotech, Fri May 10 2019, 02:41AM

I tried the photo send twice and no luck. I have the meter data and also
a page about the high voltage resistors Weston was using then.

If you log onto antiqueradios.com and ask for meter info in the test equipment
section, I will post it there . I am radiotechnician in that forum.